Hello all,
Putting on the engineer hat, here goes what I know (not much) about the 
bilinubirometers:

The spectrum of the light reflected by light changes depending on the 
concentration of bilirrubin. This makes sense because the bilirrubin 
changes the color of blood and color is related to the predominant 
wavelength of the reflected light. These devices produce a pulse of light 
and measure the spectrum of the reflected light, thus determining (based on 
calibration) the concentration of bilirrubin.

I have heard of these devices, have not seen one but I had my fellow 
engineer husband explain how they might work (my husband does research 
using optics and fluids).

I found an interesting paper comparing the use of two of these devices and 
heel prick. One of the devices cited in the paper claims to correct the 
spectrum for darker skin color.

Best regards,

Veronica Garea LLLL (and PhD, Eng Physics)
GALM Bariloche
IBFAN Bariloche - LLL Argentina



A comparison of transcutaneous bilirubinometers: SpectRx BiliCheck versus 
Minolta AirShields
C M Wong, P J E van Dijk and I A Laing
Simpson Memorial Maternity Pavilion, Lauriston Place, Edinburgh EH3 9YW, 
Scotland, UK
Archives of Disease in Childhood Fetal and Neonatal Edition 2002;87:F137-F140
http://fn.bmjjournals.com/cgi/content/full/87/2/F137

ABSTRACT
Background: Two devices are available for making transcutaneous estimates 
of serum bilirubin (SBR): the Minolta AirShields JM102 and the new SpectRx 
BiliCheck.

Objectives: (a) To measure how well the readings produced by these devices 
agree with SBR measured in the laboratory; (b) to estimate for each device, 
the proportion of infants with clinical jaundice who would require blood 
sampling if the device was used as a screening tool to detect infants with 
SBR  250 µmol/l.

Design: Prospective cohort study of jaundiced infants who required SBR 
at  20 days of postnatal age. Those who had received phototherapy or 
exchange transfusion were excluded.

Setting: Tertiary neonatal service in South-East Scotland.

Interventions: Within 30 minutes of SBR sampling, transcutaneous 
bilirubinometry was performed using one Minolta and two SpectRx devices 
(designated A and B).

Results: Sixty four neonates were enrolled, 19 of which were preterm (31–35 
weeks). The 95% confidence intervals of a device reading corresponding to 
SBR were ± 66.7, ± 67.9, and ± 66.4 µmol/l respectively. Using the devices 
to identify all SBR  250 µmol/l would reduce SBR sampling by 23%, 16%, and 
20% respectively.

Conclusions: Given that SBR levels range from 50 to 400 µmol/l in jaundiced 
infants, the 95% confidence intervals of the devices are wide at ± 67 
µmol/l. The SpectRx can be used as a screening tool for hyperbilirubinaemia 
but there is no advantage in using it over the Minolta.


-- 
No virus found in this outgoing message.
Checked by AVG Anti-Virus.
Version: 7.0.308 / Virus Database: 266.7.2 - Release Date: 11/03/05

             ***********************************************

To temporarily stop your subscription: set lactnet nomail
To start it again: set lactnet mail (or digest)
To unsubscribe: unsubscribe lactnet
All commands go to [log in to unmask]

The LACTNET mailing list is powered by L-Soft's renowned
LISTSERV(R) list management software together with L-Soft's LSMTP(R)
mailer for lightning fast mail delivery. For more information, go to:
http://www.lsoft.com/LISTSERV-powered.html

ATOM RSS1 RSS2